I am running a purity assay on a number of fractions from SMB. I am failing spec for an unknown impurity at 3.5 minutes. (%AUC =0.1%) I have eliminated the sources from the diluent and the system. It dos not appear to be a solvent. How else can I troubleshoot the identity of this unknown impurity? (No access to LCMS)
I can include chromatogram if someone will tell me how to attach it.

thanks.

Diluent: Milli Q water
Column: SymmetryShield RP18, 4.6 X 150mm, 3.5m
HPLC: 1100 w/ UV

Chromatographic Conditions:

Flow rate: 0.75 ml/min
Detection wavelength: 242 nm, BW 4 nm, Ref off.
Autosampler: 4oC
Injection volume: 5 uL (adjustable for column fractions)
Column Temperature: 45oC
Run Time: 25 minutes
Post Time: 7 minutes

Method Gradient:
Time (min) %A %B
0 70 30
5 63 37
13 63 37
22 15 85
25 15 85

What are your solvents?

If you want to post a chromatogram, make an image and use a photo upload site like flickr, and then use the Img button to display it in your post.
2 posts Page 1 of 1

Who is online

In total there are 387 users online :: 3 registered, 0 hidden and 384 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot], Majestic-12 [Bot] and 384 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry