Is there a way to assume column equivalency and substitute on column for an other.

The QA department is asking for a very valid answer with proof.

For eg:

Zorbax SBC8 4.6x150 mm 5 uM is equivalent to:

These Morons think that column equivalency means:
exactly the same dimensions (such as any C8 column with the same dimensions).

Would it help to explain that the column selection is performance based and that the vendors use proprietary processes and materials in the manufacture of the colmn packings?

It is sort of like buying a 30 year old bottle of Scotch. Basically a mixture of water and ethanol (with some traces of other stuff in there). Small differences - but a world of difference.

You may want to refer to USP Chapter <621> Chromatography. I do know there is a section in there under System Suitability and it list several changes that may be applied (for USP test methods though).

Go through the catalogs of other column suppliers - they will frequently offer direct comparisons between a given (SB or other specific type) packing and their own version (Grace, Phenomenex are examples of catalogs that do lots of this). You will probably have little trouble finding comparisons involving some at least vaguely similar molecules to what you're working with. Start there, trying out some columns of the same dimensions. If they meet your method's suitability requirements, use the heck out of a couple (from different lots) - if they perform well over the same time frame that you've been able to use the original column for, you've got it.

Fortunately for you, Zorbax SB is one packing that nearly everyone wants to offer a direct repalcement for.
Thanks,
DR
Image

How about to forward this link to your QA department:

http://www.usp.org/USPNF/columns.html

Regards

Column "equivalency" is actually a can of worms. If you have a complex separation, say an impurity profile with closely eluting analytes, your chances are slim to none to find an equivalent column. If you are doing a USP procedure for content uniformity, with one peak and one internal standard, nearly every reversed-phase column will work for you. The rest of the world is in between. The software selectivity tools (like Lloyd Snyder's tool at the USP website) can help you in selecting something that has a good chance of success, if your method is not extra-ordinarily complicated.
6 posts Page 1 of 1

Who is online

In total there are 360 users online :: 2 registered, 0 hidden and 358 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 358 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry