- Posts: 52
- Joined: Tue May 05, 2009 5:24 am
Does anyone have examples (with acceptance criteria) of system suitability specifications when testing proteins/mAbs by SEC-HPLC?
Do we inject low levels of our product (to see if the LOD/LOQ is still valid)?
Or a "mixture" which contains some covalent aggregates? Or do we re-run the MW standard curve each time, and check the plate count?
As an example, the E.P. monograph for erythropoietin specifies injecting EPO at 0.2mg/mL, and again at 0.004mg/mL (50X dilution). The peak area for the second injection should be between 1.5 - 2.5% of the first injection.
I assume this is to check that the recovery of protein from the column is ok, even at low levels...
P.S. the aim of the test is to detect HMW aggregates, or fragments (not for accurate MW sizing)