column temperature limits with Methanol organic modifier

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
My mobile phase composition is:

60% Water
40% Methanol
0.1% TFA

I am planning to use a column temp of 40C for method development purpose.

Is this temperature O.K. or is it too high in regards to the low boiling Methanol that can possibly evaporate at that temp.

It should not be a problem. The mobile phase reservoir is at ambient temperature in any case; the mobile phase doesn't come up to column temperature until it gets to the preheater tubing; since that is post-pump, the system is pressurized and boiling point is a non-issue.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanks Tom,

That was quick. I appreciate it.
3 posts Page 1 of 1

Who is online

In total there are 91 users online :: 2 registered, 0 hidden and 89 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Majestic-12 [Bot] and 89 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry