Advertisement

spiky, hairy noisy TIC chromatogram

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Greetings.

We have a Waters LC-MS.
The regular chromatogram has a perfect smooth baseline, but the TIC chromatogram is very noisy.

What could be the reason of this problem?

Thanks a lot in advance!

We figured it out.
We used methanol, 25 mM ammonium acetate as mobile phase.

We changed it to 10 mm ammonium acetate. the problem it gone.

thanks
2 posts Page 1 of 1

Who is online

In total there are 77 users online :: 1 registered, 0 hidden and 76 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 76 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry