Hello, Does anyone know how to fix the PFPD detectors?

I use PFPD with Agilent 7890B. The PFPD signal suddenly decreased. When you injected sulfur species compounds, some of the peaks disappeared and some of the peak signals are very low and broad. Like H2S and COS peak disappeared and DMS broad peak. The photometric tube works fine. The carrier flow rate Air 40 mL/min, H2 50 mL/min and make up N2 is 5mL/min. If the N2 flow is too high, the PFPD cannot be ignited.
Hello, Does anyone know how to fix the PFPD detectors?

I use PFPD with Agilent 7890B. The PFPD signal suddenly decreased. When you injected sulfur species compounds, some of the peaks disappeared and some of the peak signals are very low and broad. Like H2S and COS peak disappeared and DMS broad peak. The photometric tube works fine. The carrier flow rate Air 40 mL/min, H2 50 mL/min and make up N2 is 5mL/min. If the N2 flow is too high, the PFPD cannot be ignited.
The broad peak sounds more like inlet or column flow problems than a problem with the detector.

When was the last time inlet maintenance was performed(changing septa and liner and gold seal or swabbing out the inlet)?

Has the column been trimmed or replaced?
The past is there to guide us into the future, not to dwell in.
Hello, Does anyone know how to fix the PFPD detectors?

I use PFPD with Agilent 7890B. The PFPD signal suddenly decreased. When you injected sulfur species compounds, some of the peaks disappeared and some of the peak signals are very low and broad. Like H2S and COS peak disappeared and DMS broad peak. The photometric tube works fine. The carrier flow rate Air 40 mL/min, H2 50 mL/min and make up N2 is 5mL/min. If the N2 flow is too high, the PFPD cannot be ignited.
The broad peak sounds more like inlet or column flow problems than a problem with the detector.

When was the last time inlet maintenance was performed(changing septa and liner and gold seal or swabbing out the inlet)?

Has the column been trimmed or replaced?

Hello James,

Thank you very much for your reply!
I changed septa, liner and gold seal when the problem happened. First, I also thought that's the inlet problem. I also replaced a new column to the detector. It still has the problem. Some peaks sharp and some peaks broad. Is it the detector flow is not right?
3 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry