Residual solvent method calls for mixed standard. Response is poor when even high concentration mixed solution is injected.
However when individual solutions of each solvent are injected, response is good, even at low concentration.
What is happening?!
Residual solvent method calls for mixed standard. Response is poor when even high concentration mixed solution is injected.
However when individual solutions of each solvent are injected, response is good, even at low concentration.
What is happening?!
What solvent is used for each individual analyte when injecting them separately? Are they being injected neat or as a dilution separately and in mixture?
The past is there to guide us into the future, not to dwell in.
Residual solvent method calls for mixed standard. Response is poor when even high concentration mixed solution is injected.
However when individual solutions of each solvent are injected, response is good, even at low concentration.
What is happening?!
What solvent is used for each individual analyte when injecting them separately? Are they being injected neat or as a dilution separately and in mixture?
The diluent for all injections is 90:10 1,3-Dimethyl-2-imidazolidinone:water.
I've injected all as a mixture as per the method,and each separately in diluent at LOQ concentration.
LOQ injections have higher responses than the highest concentration mixed solution .
Residual solvent method calls for mixed standard. Response is poor when even high concentration mixed solution is injected.
However when individual solutions of each solvent are injected, response is good, even at low concentration.
What is happening?!
What solvent is used for each individual analyte when injecting them separately? Are they being injected neat or as a dilution separately and in mixture?
The diluent for all injections is 90:10 1,3-Dimethyl-2-imidazolidinone:water.
I've injected all as a mixture as per the method,and each separately in diluent at LOQ concentration.
LOQ injections have higher responses than the highest concentration mixed solution .
What are the stock concentration of each before mixing and how are they mixed? What should the final concentration of the mixed solution be?
The past is there to guide us into the future, not to dwell in.
4 posts Page 1 of 1

Who is online

In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry