we are not getting consistant areas of n-propnaol by headpace , what could be the reason, column zb-624, 30mx0.53mm, 3.0µm, please suggest
thanks
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Chromatographic Parameters:Hi,
Please give us more info:
Which incubation temperature and time are you using?At what temperature is the inlet and detector?Are you working isothermal?Which liner type are you using? How much %RSD are you getting?
Regards,
Atom
i tried with Oven temperature 80°C and noticed inconsistant areas and then moved to 100°CRewriting due to late post.......
My first guess:
a. 100°C in HS oven when using DMF (bp ~155°C) gives you a load of DMF entering the valve and load. Personally avoid using more than 70°C if possible. This in combination with b.....
b. Temperatures in valve/loop and transferline is below DMF bp. Increase to above bp of DMF.
Polar compounds like propanol can attach to residual/condensed DMF in valve/loop => inconsitant areas
c. loop fill time might be a bit long but likely not, never had to go outside the 0,2-0,4min range with similar applications.
d. Even though DMF retains solvents better than water, the high HS temp of 100°C compared to bp of propanol may contribute to the high RSD%, if your sample solvent had been water it most likely would have.
You should not need that high oven temp 60-70 °C would surely be enough.
Well that increase both amount of DMF and propanol entering the loop. It does not prevent DMF from condensating in the valve/loop, rather increasing the chance of contaminating surfaces in valve/loop with DMF.
i tried with Oven temperature 80°C and noticed inconsistant areas and then moved to 100°C