I am currently using a Tekmar LSC2000 P&T in combination with a HP 5890 GC. The current setup has the deactivated column from the heated line directly passing through the septum of the GC injection port. As it stands there isn't a leak showing up in the results but I feel the setup looks a little precarious to work as a permanent design. Has anyone had much experience with the P&T and GC combination and can recommended a "better" arrangement? Thanks for any help in advance.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Have you searched the forum ? - there are dozens or hundreds of posts on P&T-GC
Peter
Peter
Peter Apps
-
- Posts: 33
- Joined: Wed Mar 18, 2009 12:15 am
How is the transfer line end secured on the injection port? Are you saying that the tip of the transfer line is punctured through the septum?
If so and you have any say in the matter, you can order an injection port weldment adapted for a P&T transfer line for about $300-400 from Restek.
If so and you have any say in the matter, you can order an injection port weldment adapted for a P&T transfer line for about $300-400 from Restek.
-
- Posts: 199
- Joined: Mon Sep 22, 2008 7:50 pm
You really have two options for the carrier, depends on the design of the Tekmar. One is to run all the carrier through the HS, the second is to run a partial carrier stream through the HS. In my experience, the first option is more common and easier to understand. We had a GC here that someone had set up with a small bore carrier line shoved through the septa (it was for an LSV). It's worked for years that way. Never understood why they didn't just cut the carrier line and install the LSV there.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
The major disadvantage of plumbing a purge and trap or headspacer transfer line into the carrier gas feed line is that it exposes the sample to a length of undeactivated, unheated metal pipe.
Peter
Peter
Peter Apps
-
- Posts: 119
- Joined: Mon Nov 26, 2007 5:46 pm
Back in January, in the GC area, under the heading "Mass Flow /Back Pressure" by "pueblito" I responded with a diagram for configuring a tekmar P&T to a 5890. Assuming I can paste a link that you can follow:
viewtopic.php?t=9960&start=0&postdays=0 ... ight=purge
If you have any questions, I'll be checking back.
Greg
viewtopic.php?t=9960&start=0&postdays=0 ... ight=purge
If you have any questions, I'll be checking back.
Greg
6 posts
Page 1 of 1
Who is online
In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science