shoulder peak problems, overlapping peak problems
Here's the situation:
We spiked hydrocortisone and prednisone (IS) in urine. We used 40:60 ACN:H2O for our mobile phase and a C18 column. Flow rate is 0.70 mL/min. After adjusting pH to 9.0 and increasing ionic strength, we extracted analytes using solvent extraction. Our first run, we noticed in the chromatogram that a peak appears very closely to the hyrdocortisone peak. After increasing the hydrocortisone concentration, we noticed that there appears to be overlapping of peaks. Any suggestions on how to separate them?
I know this should be easy...but we are rushing. We need solutions ASAP. Decreasing the flow rate will probably separate the peaks a little but it will not completely eliminate the problem. A relatively high concentration of hydrocortisone (greater than 10 ppm) will most likely end up having a shoulder peak because of the overlapping peaks.
You can:
a. decrease particle size from 5um to 3um
b. increase column length
c. replace ACN with MeOH or THF
Below is data showing this separation on Cadenza CD-C18 and Cadenza CL-C18:
http://www.imtakt.com/TecInfo/TI039E.pdf
http://www.imtakt.com/TecInfo/TI324E.pdf
Check out a column with an embedded polar group (SymmetryShield RP18, XBridge Shield RP18). Such columns give a different selectivity for analytes with a phenol function.
How about the most obvious: fiddle with the H2O/org ratio? Or even before that, check the lit., there should be a trillion examples.
You can:
ok ill fo with increasing the column length first...