I am working on perkin elmer GC FID with total chrom software. I need to develope a method for TPH analysis using "timed group". Can any help me?
Posted: Mon Mar 30, 2009 2:34 pm
by DR
Timed group is an "on/off" function where you assign any peak within the window to a group ID, then they're calculated as a single peak using the sum of peak areas within the window (if memory serves correctly).
empty report
Posted: Mon Mar 30, 2009 3:51 pm
by chemjeeorganic
Thanks for the reply. I segregated the diesal range into two groups one >C10 to <C18 and second group is >C19 to <C29. and they were assigned in the time group wind but when run was done in the final report it is said that there is no group to be identified.
I run 1000ppm diesel standard.I am getting very good chromatogram. it seems i an going some work wrong inthe method.Can you pls tell in detail how to enter the time group in detail?
Thanks
NJ
Posted: Tue Mar 31, 2009 3:54 pm
by DR
I can't (it has been several years since I used T*C now), but I suspect someone else can, or you could check the Users Group at the vendor's website.
Forum for users with service plans
T CHROM
Posted: Wed Apr 01, 2009 3:37 pm
by chemjeeorganic
Sir
Many Thanks for the information.
Regards
NJ