UPLC baseline drift

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I am currently optimising a UPLC purity/content method for a PEGylated peptide drug product. To elute small N-terminal fragments, the gradient at the front was adapted, starting from a lower percentage of acetonitrile (0.5% instead of 5%). The slope remained the same.

The following set-up is used:
-Waters Acquity UPLC H-Class with 425 µL mixer
-Flow rate: 0.30 mL/min
-Column: Waters Acquity CSH C18, 130A, 1.7µm, 2.1 x 100mm
-Detector: UV or PDA at 215 nm
-Mobile phase A: 0.05% TFA in water (high grade TFA is used in 1mL ampoules)
-Mobile phase B: 0.04% TFA in ACN


Using this gradient on a degraded drug product sample containing the N-terminal fragments, we get a desirable separation of the degradants, however the baseline is not flat.

Although we are already using a larger mixer, we hypothesised the mixing of the mobile phases had something to do with it. Therefore, we prepared a premixed mobile phase with 99.5% H2O and 0.5% ACN (with 0.05% TFA).
Using the premixed mobile phase, we already get a better result, but still the baseline is not really under control.

We additionally tested:
- other supplier of TFA: no influence
- slower gradient (50% slower): no improvement
- longer equilibration at 0.5% ACN at start (2.5min instead of 0.5min): no improvement
- mobile phase B (ACN) without TFA: same 'jump' visible

The same method was run in a different lab on Agilent 1290 UPLC system and there the baseline was much better.


Is this something system related? What could possible solutions be? As we are working in a GMP context we do not have much room to change the set-up of our equipment unfortunately.

Thanks a lot in advance!

all relevant images:


Image

EDIT: embedding image did not work, this is the link to the images:

https://imgur.com/a/YlUTfn3
Dear you are getting the drift or ripples please confirm.
If you are getting the ripples please revert so that i can share my experiance as i faced ripples problem in UPLC which is related to agilent and i also found the solution of that ripples after almost 9 month working on it.
Hi, our issue is indeed the ripples in the baseline. We are e however using waters h class uplc. I am looking forward to hear your thoughts about it!

KR

Willem
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 2 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], John Guajardo and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry