ksharp » Fri Mar 27, 2009 4:54 pm
As far as I can tell, these are the only tests that really need to be performed:
Auto-Sampler
-Flow Rate Test {1.0 ml/min @ a known wavelengths. Other flow rates can also be checked if need be: 1.5 ml/min, etc.) Note: All solvent lines need to be checked.
I don't think you need a specified/known wavelength to measure flow rate, you don't even need your detector on (please correct me if I am wrong). Test flow rate with water as mobile phase; collect mobile phase in a weighed volumetric flask for exactly ten minutes, as it leaves the system (after the column) and determine the difference.
-Precision or Reproducibility
Injector precision is verified by %RSD over six injections.
-Carryover (is this required or necessary?)
Yes. Inject a sample with high concentration of analyte, then perform injection containing zero analyte but in the same solvent. Area/amount in the second sample relative to the first is the carryover.
Detector: Wavelength Accuracy
Agilent sells certified caffeine reference standards, local minima and maxima are measured on your instrument and compared with known values (244 nm min., 272 nm max.)
Column Oven: Temperature Accuracy
We measure with a external thermometer (NIST traceable), set instrument to a known temperature and measure it, compare that with the instrument setpoint.