Thank you for the details. That gives a much clearer picture of the situation.
Now I can't really tell what it is but some more suggestions:
- if all the pressure and leak tests pass, then it would be weird that air should come from outside... but one never knows...
- if you immerse the out of the detector in a liquid, do you see an air-bubble when the peak tails?
- Is your degasser working properly?
If I remeber, the pressure indicator should be below 40 psi (or was it mbar)?
- do you have a pressure regulator at the out-port of the detector, so it won't siphoning or bubbles in the cell?
If you use a small piece of tubing, make sure it won't exceed the cell's pressure limit under your working conditions
I think something similar was seen in another thread here in the forum, with bubbles/weird baselines occuring once in a while after several good injections.
maybe with a real chromatogramm with several substances, one could see, if all peak are affected (real injection problem, void volume) or only the first ones ->extra column effects or timed/volumed issues.
-> LC Resources Troubleshooting Wizard:
http://www.lcresources.com/resources/TSWiz/hs40.htm