Bibenzyl extraction -- Solid phase extraction
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
I am currently doing a project, to detect bibenzyl, dibenzyl disulfide in insulating oil. We are trying to use 500mg alumina neutral to extract both from insulating oil. However, I only manage to extract dibenzyl disulfide. It seems that bibenzyl passes through the catridge w/o being absorbed. Anything wrong with the SPE selection?
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
laucch123,
How do you know it is not being adsorbed versus it is not coming off the cartridge? Can you tell us more?
Best regards.
How do you know it is not being adsorbed versus it is not coming off the cartridge? Can you tell us more?
Best regards.
-
- Posts: 247
- Joined: Sat Apr 28, 2007 3:24 pm
It will only be absorbed if the elution solvent is weaker than the analyte. In normal phase chromatography (ie alumina), less polar solvents are weaker than more polar solvents. So bibenzyl will only sorb if your solvent is hexane (or heptane, etc). It can then be eluted using pretty much any other solvent. If you give some more details of your procedure, maybe we can be more help.
-
- Posts: 3
- Joined: Wed Mar 18, 2009 7:36 am
I had prepared these two components in an insulating oil. Then I take 500mg of this calibration standard diluted with 1cc hexane, then the diluted standard penetrated into the SPE catridge. The catridge is then washed with 1cc hexane for three times. Finally with 1cc of benzene. And analyze the benzene with GC MS. I also analyze the hexane eluted from the catridge and found that most of the bibenzyl is in there. I think this will work but now just no idea what is freaking wrong..
-
- Posts: 247
- Joined: Sat Apr 28, 2007 3:24 pm
Perhaps their are aromatics in the oil that are saturating the SPE cartridge and causing the bibenzyl to elute in the hexane fraction. Maybe you can try using a smaller volume of oil or diluting the 500mg to a larger volume prior to loading it onto the cartridge?
5 posts
Page 1 of 1
Who is online
In total there are 38 users online :: 1 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 37 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 37 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science