- Posts: 5
- Joined: Fri Jan 31, 2020 6:05 pm
I have a real problem in obviating noises and drifts I see in the baseline chromatogram and even when no injection is running and lines are simply getting filled with solvents (No column, Union Attached).
I'm working with an Agilent 1100 equipped with a DAD detector. I have checked following:
Pump is working fine, the flow pressure when a phosphoric buffer is running off the column is at most at 100 bar. When purge is open, the pressure wouldn't do upper than 10bar. Also, with union attached and purge closed I haven't observed pressure more than 30 bar. Flow of the pump is true (I checked from 0.5mL/min to 5mL/min and each time the flow was correct)
PTFE filters behind the pump is tidy and unsaturated. Also, active inlet is fine.
The column is Siliachrome from Silicycle (25cm, 6 micron, 4.6mm) and the mobile phase has buffer in itself. I see too much noises when it is attached or when its not (no difference).
I have done self test in the instrument, holmium test and other tests separately and they have passed successfully.
Plus, I have passed following solvents in three days, each for 12 hours with no success (because I think flow cell must be contaminated which should be cleared with the following solvents but nothing good happened unfortunately).
- Isopropyl alcohol
- Acetonitrile
- Water
- Methanol
Plus, I increased the Peakwidth and Slit from 2s & 4nm to 16s & 16nm (in this case, noises were smoothed but I know it is still there. Also, the negative drift decreased from +-5mAU to +-1mAU).
The last try was to check with different wavelengths as follows (For Peak-width of 2s and Slit of 4nm):
210nm, BW: 16nm- Ref 360nm, BW: 100nm. Result: Negative drift +-5mAU, Noise of 5mAU
230nm, BW: 16nm- Ref 360nm, BW: 100nm. Result: Negative drift +-3mAU, Noise of 5mAU
260nm, BW: 16nm- Ref 360nm, BW: 100nm. Result: Negative drift +-3mAU, Noise of 3mAU
286nm, BW: 16nm- Ref 360nm, BW: 100nm. Result: Negative drift +-1mAU, Noise of 2mAU
Is it acceptable to have a 5mAU of noises and Negative drift of +-5mAU at 210nm when Peakwidth and Slit are 2s and 4nm, respectively? If not, what should I do more to decrese noises and drifts?
(Lab temperature alternates but it is mostly at 22 degree celcius. Plus, instruments' electrical plugs are attached to an adapter which compensates for alterations of electrical voltage. For the last thing, detector lamp has worked for roughly 1800 hours)
Thank you all for your helps.