GC/MS
Posted: Tue Mar 10, 2009 3:38 pm
We are using a 6890 with a 5975 MSD from Agilent. The foreline pump had trouble some time ago, we disassembled it and it functioned for about one month more. Last week the pump failed and I found that the oil level was down. I disassembled the pump again and put more oil. The pump apparently works ok because the pressure in the analyzer is 4.01e-5. But when we try to do an autotune, the equipment stops, saying that peakwidth exceeds the limits. Also you see only noise in the profile of the tune.
I am checking with manual tune masses 69, 18 and 28 doing a scan. I am monitoring the relationship between 18/69 and 28/69. This are in the order of 521 and 931 % after two hours of pumpdown start. I have several times checked for leaks in the system, but cannot find one that should be very large for the amounts stated. I think that the problem is the pump, what do you think? Somebody knows another form of testing it?
The vacuum system is a standard turbo pump.
Thank you,
Adolfo
Thank you,
Adolfo