Hi,
I have cleaned the ion source. With the procedure that recommend Agilent. They recommend first to
1 Disassemble the ion source.
2 Collect the following parts to be cleaned:
• Repeller
• Interface socket
• Source body
• Drawout plate
• Drawout cylinder
• Ion focus lens
• Entrance lens
These are the parts that contact the sample or ion beam. The other parts
normally should not require cleaning.
3 If the contamination is serious, such as an oil backflow into the analyzer,
seriously consider replacing the contaminated parts.
4 Abrasively clean the surfaces that contact the sample or ion beam.
Use an abrasive slurry of alumina powder and reagent-grade methanol on a
cotton swab. Use enough force to remove all discolorations. Polishing the
parts is not necessary; small scratches will not harm performance. Also
abrasively clean the discolorations where electrons from the filaments
enter the source body.
5 Rinse away all abrasive residue with reagent-grade methanol.
Make sure all abrasive residue is rinsed way before ultrasonic cleaning. If
the methanol becomes cloudy or contains visible particles, rinse again.
6 Separate the parts that were abrasively cleaned from the parts that were
not abrasively cleaned.
7 Ultrasonically clean the parts (each group separately) for 15 minutes in
each of the following solvents:
• Methylene chloride (reagent-grade)
• Acetone (reagent-grade)
• Methanol (reagent-grade)
I have done the steps above took from the manual. When I cleaned with Methylene chloride it turned green. After the wash in ultrasound bath the pieces looked with spots and stains, so I cleaned again with same procedure and the stains remained there. So I looked after other methods and used one that found in
www.sisweb.com. They used a mild detergent, water, acetone and methanol to wash the pieces. The results were better. So after the cleaning I assembled the source again. Wait for the stabilization and after some hours (5). I tried the autotune. The surprise is that the filaments were out. In the 5975 you can see the filament light when they are on through a glass window. But this time there was no light. So I vented again and checked with ohmeter filaments, etc. on the source. I assembled again and this time I tried the filaments only 1.5 hours after pumpdown complete. The filamnets went on. I let the equipment then four hours more stabilizing but when I tried to tune, the filaments doesn't light again. After some more intents, the massspec doesn't show any peaks and only shows a message telling that the MSD has not found any peak.
Do you have some ideas.
Thank you for your help and time.
Regards,
Adolfo