We've got another issue I would like to ask for inputs.
We're trying to establish catecholamine analytics with ion pair reagent on C18 column, as published in varius application notes.
After the first few trials which looked very promising, the retention time of dopamine decreases with about 0.5 min for every injection, starting at about 22 min.
Does anybody have an explanation of this phenomenon and probably also have a solution on how to solve it?
I searched the old forum archive and saw, that similar issues were observed some time ago but didn't find a solution.
The decrease shows good correlation with almost every parameter, time, no of injections or volume of eluent (maybe we didn't too few variation)
I don't think it's due to incomplete column equilibration, because the retention time doesn't stabilize even after >100 column volumes of eluent pumped. Also the column is thermostated at 40°C, so thermal fluctuations are not likely.
The "fault" we've done is to flush the column with 100% MeOH to clean the column from unpolar sample compounds and for storage the column. Then after a week and reequilibrating with the ion pair reagent, we got these drifting retention times...

our chromatographic system:
Col: Symmetry 4.6x75mm, 3.5µm
Eluent: Na-Octansulfonate 5 mM/PO4-buffer 85mM, pH3/MeOH 5%, isocratic
Flow: 1 ml/min
Temp: 40°C
Pressure: about 2700 psi
Inj Vol: 10-20 µl
sample: dissolved in 0.1 M HCl, same effect if diluted or dissolved in buffer.
Areas are not affected by the drift.
Any suggestions?
Could it be possible to be some "dewetting" effect due to the high water content on the C18 column? (>we have to do some experiments on this)
Are we disturbing the ion pair equilibrium with each injections? (> but longer cycle times doesn't improve/restore the retention time)
PS: before we used ion pair we already tried:
- conventionell reversed phase C18 on Atlantis T3 with 100% water
>good selectivity with standards but not with real sample (matrix)
-- SPE clean up with ion exchange didn't solve the problem.
- Atlantis HILIC: no satisfactory separation/reproducibility/peak shape