Is it possible to see formaldehyde by gc ms without derivatization?
If it is, what are the conditions (columns, etc)?

I don't know if by MS but it is doable by GC with the right detectors. I will look in my notes to see what column was used.

Best regards.

Thanks for the reply. GC alone is fine, as long as formaldehyde doesn't need to be derivatized.
Also, how can one obtain formaldehyde in a gaseous state? What kind of containers are used?

Thanks.


Also, how can one obtain formaldehyde in a gaseous state? What kind of containers are used?

Thanks.
Hi

Have limited experiance of this but seems to recall from the university that we used a diffusion sampler (passive sampling) imprignated with 2,4-dinitrophenylhydrazine. The reaction product was analyzed with HPLC.

Sorry have no further details

berettagtz,

Still don't know what column but I will continue looking. Regarding formaldehyde in gaseous state, I would start with the permeation tube manufacturers like Kin-Tek or Valco. You can also make it from paraformaldehyde once you heat it. Controlling the concentration will be the challenge in that scenario. If I am not mistaken, I think someone made a perm tube of paraformaldehyde and controlled the concentration by how much they heated it.

More to come.

Best regards.

Thank you for your replies. So far I haven't found any company that sells formaldehyde in a pure gaseous state. As far as the column goes I may have found a GC column from sigma aldrich.

Thank you for your replies. So far I haven't found any company that sells formaldehyde in a pure gaseous state. As far as the column goes I may have found a GC column from sigma aldrich.
Try Scotty and Entech

Thank you for your replies. So far I haven't found any company that sells formaldehyde in a pure gaseous state. As far as the column goes I may have found a GC column from sigma aldrich.
Try Scotty and Entech
Would you happen to have a website?

Google search find following links:

http://www.scottgas.com/

http://www.entechinst.com/

berettagtz,

I found the reference. It was a poster presented at PittCon by Air liquide. They used a standard in nitrogen that they had prepared at about 20 ppm. they used a 0.53 id RT-QPlot column, 30 meters long with an Agilent GC-Ms. I am not sure how they married 0.53 flow to a MS since they don't say anything in the poster. They also did it by GC HID with a 10M mol-sieve and a 50 mtr PorabondQ column but they don't discuss how configured. For this, they used a perm device from "Calibrage?" (I am assuming a French perm tube maker....)

Best regards.
10 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry