Re: On-Column - a possible misunderstanding of the basics

Posted: Wed Jul 04, 2012 12:39 pm
by mbdixon
The Carrier gas flow rate is 3ml/min (Nitrogen). Most things I have read suggests a slightly higher than normal flow rate for this sort of analysis.

Re: On-Column - a possible misunderstanding of the basics

Posted: Wed Jul 04, 2012 12:54 pm
by Peter Apps
The Carrier gas flow rate is 3ml/min (Nitrogen). Most things I have read suggests a slightly higher than normal flow rate for this sort of analysis.
OK, I would agree with that.

With the column at 50C there is a danger that water will condense in the column, which will distort the peaks and probably damage the stationary phase.

What concentration of analytes do you have in the samples ? If the concentration is high enough you could try a split injection, which makes the water matrix a lot more tractable. Trouble is that Uniliners with the hole at the top do not owrk well with splitting.

This is a resurrected thread started off by someone else - can you summarize your problem and set up so that everyone is sure what it is that we are talking about.

Peter

Re: On-Column - a possible misunderstanding of the basics

Posted: Wed Jul 04, 2012 1:11 pm
by mbdixon
Cheers Peter, I've upped the initial temperature to 80°c and opened a new thread. Thanks