AICMM,
Yes, I was having good (better) luck with this until I went to 80°C. The compounds I was working on were sort of a pre-updated <467> test for some of our clients. Since we now have the new <467> to work from, I am modifying our method to analyze for the Class 1, 2A and 2B categories. I am not concerned with the delay of the <467> update, since we can offer the analysis to those clients that want it, instead of the normal OVI analysis. So, the 80°C came from the updated <467> and I think a deviation back to 60 would probably put any results into question, because of insufficient heating.
For our clients that want to continue with our current residual solvent profile, I want to improve the peak shapes and reproducibility I am seeing, while moving towards the method modifications necessary to meet the new <467>.
The current sweep time is 2 minutes. The longer the sweep, the further the molecule will travel and the more water will accumulate. Reducing this is next on my list of things to try.
The last chromatogram on the previous page is from a trap analysis. I only mention the headspace analysis in passing. I will again try static sampling with the analytes in our current profile (since the DMF and DMSO have been excluded).
Rod,
That is another good idea with an additional portion of a -wax column. However I would be concerned about the possibility of air interfering with my MS. Is there a good solution to glass press-fit connections? I would assume that I should start with ~1m and increase if necessary. Or should I start with 5m right away?
I also think I forgot to mention that the above chromats (one marginal, one poor) are from the same sequence, just separated by 10 samples.
Thanks,
Schmitty