I'm with Peter on this one, if there is junk at the end of the run, it could be eluting just after the end of data collection on the working instruments and just being trapped by the cooling oven on the non working instruments.Then your interference may be a ghost peak. Is it a problem to simply extend the run time ?
Peter
If that is not the case, the only thing I haven't seen mentioned is the gas supply tubing being changed. You mentioned changing gas traps, but if some contaminate got into the tubing between the trap and instrument, then it could be contaminating the inlet. If it only appears after a sample injection, then maybe something in the sample matrix is knocking it loose when just the solvent in the standards does not. That is a long shot, but can be considered if extending run time doesn't fix the problem.