Posted: Mon Apr 10, 2006 10:24 pm
Thank you Mark for your reply.
I am using the system for more than two years in the same lab and location and it was working fine for that period of time. This is why I was eliminating any effect of airconditioning. I also do not use any online degassers for this system. I degass the mobile phase by sonication.
For these tests I replaced the column with a piece of tubing to get enogh pressure to operate the pump. However, the Noisy_baseline_4 was generated at really low pressure approaching the minimum limit.
At higher flow rates (0.5, 1.0 and 2 ml/min) I get a similar baseline to what it is shown in Noisy_baseline_3.