Re: GC/FID TPH (C10-C40) Baseline Shift Problem
Posted: Fri Sep 02, 2016 7:43 pm
Hi, Qazedc;
I think Peter & Hornet are right on the money - the first thing to do is change that 6-year old trap.
If you get oxygen inside the column, then elevate the column oven temperature like you're doing each run, you will definitely badly damage the column phase, resulting in high bleed.
Try that first, see how an injection looks after that, but you might just be stuck replacing the column anyhow if the phase is already damaged.
Best of luck!
Donna