Page 2 of 2

Re: Rising Baseline in Gas Chromatography

Posted: Mon Jan 21, 2013 9:01 am
by Peter Apps
I agree with Don, the baseline drift is acceptable for the size of peaks that you have, provided that the integration puts the baseline in the right place as in Fig 4. Before you start swapping ferrules and trimming the column you need to check for leaks - if there is no leak there is no point swapping ferrules (do not fix what ain't broke rule), if there is a leak you must fix it before you trim the column. Check the septum also.

Peter

Re: Rising Baseline in Gas Chromatography

Posted: Mon Jan 21, 2013 5:22 pm
by kristind858
I disconnected the column from the detector and inserted a no-hole ferrule. I ran my method and this is the chromatogram I got:


Image


Figure 6: Blank run-No injection-Nocolumn connected to detector. Temperature program: 170 deg C for 0 min, then 0.6 degC/min to 207.5 deg C for 0 min, then 5 deg C/min to 225 deg C for 0 min. Column: ZB wax 100% PEG.


So I guess it was the column and the detector does not need to be cleaned.

Re: Rising Baseline in Gas Chromatography

Posted: Mon Jan 21, 2013 5:24 pm
by kristind858
Thanks Peter and Don !!!