The Carrier gas flow rate is 3ml/min (Nitrogen). Most things I have read suggests a slightly higher than normal flow rate for this sort of analysis.
The Carrier gas flow rate is 3ml/min (Nitrogen). Most things I have read suggests a slightly higher than normal flow rate for this sort of analysis.
OK, I would agree with that.

With the column at 50C there is a danger that water will condense in the column, which will distort the peaks and probably damage the stationary phase.

What concentration of analytes do you have in the samples ? If the concentration is high enough you could try a split injection, which makes the water matrix a lot more tractable. Trouble is that Uniliners with the hole at the top do not owrk well with splitting.

This is a resurrected thread started off by someone else - can you summarize your problem and set up so that everyone is sure what it is that we are talking about.

Peter
Peter Apps
Cheers Peter, I've upped the initial temperature to 80°c and opened a new thread. Thanks
33 posts Page 3 of 3

Who is online

In total there are 50 users online :: 1 registered, 0 hidden and 49 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot] and 49 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry