getting a SRI model 8610 GC to talk to the Win 95 computer
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
3 posts Page 1 of 1
I am trying to get a older SRI GC Model 8610 (02/28/1995)back on line. We are having difficulty getting the GC to talk to the Win 95 computer thru the computer’s serial port. When we start up the software we get "Can not open serial port error" and/or "Acquistion system is not functioning" and/or "Can't wake up check power & cable."
Peak Simple version 2.83, Pentium computer 192 MB ram 1.5 GB hard drive.
Any ideas?
Tom McCall
reply to this post or to my email - coastalva@yahoo.com
Experienced Analytical Chemist / Instrument Service Engineer -
http://www.linkedin.com/pub/tom-mccall/a/2a0/938
you know software is free? You can just download it from here.
http://www.srigc.com/PeakSimple.htm
So try it on a newer computer with a serial port and see if this corrects the issue.
Make sure you have the right com port selected in your options. Edit-> Overall -> change the com number to the one you have your gc connected too. I have a SRI 8610C connected to a win 95 computer, so the operating systems is not your problem.
3 posts Page 1 of 1
In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)