Advertisement

What mechanism can cause this?

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I made a 50% toluene 50% tetradecane by weight standard, mixed, and got 27% tol and 73% tetradecane by Area% on my 5890 Series II TCD detector!!! I shot the same mixture on a 6890 with same conditions and got 52% toluene and 48% tetradecane by Area% on its TCD (which is what I expected). Has anyone ever seen this type of discrepancy or have any idea what is wrong with the 5890?

Thanx for your time.

My money would be on inlet discrimination.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 71 users online :: 1 registered, 0 hidden and 70 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 70 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry