we actually use a bubble flowmeter. how close do you think this is compared to a column with no bubble flowmeter attached?
I cannot think of any way of accurately volumetrically measuring a volume flow of 3 ml/min through a column that is connected to a detector with the detector gasses flowing to match the back pressure that you get when running an analysis. This is the advantage of measuring linear flow rate in terms of the retention time of an unretained compound - you can do it with the column installed to the detector and under exactly the same conditions as you use for the actual analysis.
Peter