Harvard PHD 22/2000 pump
Does anyone use Harvard PHD 22/2000 pump for post column derivatization usage? How about the pulse during the flow switching? Can it be eliminated by adding dampener?
Thank you.
Syringe pumps do not need pulse dampers. In fact, pulse dampers require at least 500 psi of back pressure before they do anything at all, and few syringe pumps have both the pressure rating and volume capacity to operate in that range. (That is why the Pickering syringe pump is so costly, it is special-made for that operating region.) For your application, you do not need a long reaction loop, nor high temperatures, so the operating pressure will be very low.
What you will want to do is to have some relay controls to refill the syringe at the beginning of each analysis. Then start the syringe pump, wait for a stable baseline, and start the chromatography.
Mark Tracy
Senior Chemist
Dionex Corp.
In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 23 guests
