- Posts: 73
- Joined: Tue Jul 04, 2006 8:33 am
Good luck,
Salma
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Hi Michelle,'''
Syringe Size = Glass 1ul
Sample size injected = 0.1ul
chemical components =
Butan-1-ol, 116 -118C boiling point (bp), Retention time (RT) = 3.78min, Area = 348877
IBMK (internal Standards), 117-118C bp, RT = 4.94mins, Area = 349099
N-butyl acetate, 126-127C, RT = 7.66mins, Area = 317879
Best GC conditions so far
Oven Temperature (Temp) = 45C Isothermally
Injection Temp = 135C
Detector Temp = 170C
A tiny bit of washing-up liquid in a beaker of about 10ml water. Pour about 1ml into rubber squichy bulb thing at the bottom and attach to the end of the glass tube. There should be a long thin plastic tube attached to the glass tube just above the rubber bulb. Attach the plastic tube to the vent and press the rubber bulb a couple of times until a bubble travels above the entrance of the plastic tube to the glass tube. The gas traveling along the plastic tube from the vent will push the bubble up the tube and when it passes the 0 mark press go on the timer until it reaches the graduated 10ml mark and then stop. thanks Bruce" how do you get this quote thing to work?!!!
MakeupHi,
Syringe
4 Segment syringe technique.
Don't understand this technique at all. Sorry
Syringe
I have a 1ul syringe and I am trying to inject 0.1ul of sample.
Make-up gas flow
I set the make-up gass flow to the same as the carrier gas flow. I have this on all the time. I found this to make a difference to the peak shape. What does this do. I know the carrier is going through the column but the make-up???/
I should have commented on this. Firstly, always try to relate flows to mls/min. Using that unit will help you visualise flows, eg..Purge & Split Vent flows
Purge Vent Flow: I opened this valve and set the flow to around 2ml/min or 5sec/ 10ml
Split Vent flow: Measured this again. It is giving conflicting results. So I definetly set it this time to around 150ml/min or to be exact 3.6sec/10ml
If the method states that the split flow should be 1:50, does this mean 1ml in 50min or is it 50ml in 1min. I am getting confused with the units.
Should I adjust the split flow to 1:50 and will that allow me to operate at the higher temperature of 60C as stated in the method or should I just keep it at the 150ml/min
If your samples are fairly clean, then the 624 column will separate the compoents. It's worth noting that a polar column like 624 will bleed stationary pahse much more than your methyl silicone, so can't be taking to such high temperatures, and is more sensitive to oxygen.Hi All,
What column should I choose to achieve this separation?
I think that sticking with the 30m and maybe a 0.3um diameter??
What film thickness? 1.4um would 1.0um or 1.3um be ok??
Someone mentioned to me that I should set the detector and Injector temperatures to the following values respectively: 300C and 200C (min.) This is to ensure that all components are completely vaporised.
"I am using an Elite-624 capillary column (30m * 250um * 1.4um film thickness) and I can separate these fine"
Thanks
Michelle