Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

18 posts Page 2 of 2

Sluggo, the retention time behavior may be a function of the compound pKa. If the mobile phase pH is close to the pKa of the compound you are analyzing, small changes in pH can have dramatic influence on selectivity and retention time. Do you know what the pKa of your compound is?

i've never observed this kind of variation with 0.05 change of pH scale
i suggest this:
Since i think your compound is completely protoned at 2.4
the error is in my opinion the second readjust of pH, i never do it or see it done.
tea can be useful but there are a lot of vendor who sells column with good perfrormances solving the tailing problem without using tea (you may try a column with a high surface area like symmetry or synergi playing with the strong solvent in case of too much long retention times).
vincenzo

It now looks like the issue was due to high sensitivity to organic %.

I am now using a different method for this compound.

thanks
18 posts Page 2 of 2

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry