Thanks to those who replied on my query regarding cefuroxime stability. I have a new problem though:

1. We run system suitability first before proceeding with any validation run and in the result, the RT of cefuroxime is repeatable with RSD of less than 2.
2. During the validation run, we noticed a change in RT for both cefuroxime and IS sulfamethoxazole.
3. RT changed from 6.2 to 6.6 minutes for Cefuroxime and 7.5 to 8.0 minutes for sulfamethoxazole.

M.P.: MeOH/ 0.005 M KH2PO4 buffer (pH 3.6), 30:70
Sample treament: Analytes are in plasma, precipitated with Acetonitrile and back-extracted with DCM/Hexane (1:1)
Stock and Working Solutions are dissolved in 0.01M KH2PO4 buffer, pH 6.5

Is this change in RT significant? Is there any part in the sample preparation or Mobile Phase that can influence the change in RT?

I will appreciate any feedback!

Thank you.
Ghie Malig :wink:

Hello :D

Several factors can influence RT : flow rate, mp composition, column chemistry, temperature, injection technique, to name some. Sample prep won't usually affect RT significantly, unless your sample prep method is either extracting the wrong analyte or is changing the analyte structurally, (usually because of changes in pH )

However, assuming this is not happening and since you're using an internal standard, keep an eye on the relative retention time of your analyte w.r.t. the IS. If the RRT is repeatable within limits of RSD and if your system suitability is otherwise OK, don't worry too much about the occasional shift in the analyte's uncorrected RT - provided of course, the relative retention time remains constant.

Hope this is useful.
SK Srinivas, MPharm
CEO, K-Prime
Chromatography Training
Thanks to those who replied on my query regarding cefuroxime stability. I have a new problem though:

1. We run system suitability first before proceeding with any validation run and in the result, the RT of cefuroxime is repeatable with RSD of less than 2.
2. During the validation run, we noticed a change in RT for both cefuroxime and IS sulfamethoxazole.
3. RT changed from 6.2 to 6.6 minutes for Cefuroxime and 7.5 to 8.0 minutes for sulfamethoxazole.

M.P.: MeOH/ 0.005 M KH2PO4 buffer (pH 3.6), 30:70
Sample treament: Analytes are in plasma, precipitated with Acetonitrile and back-extracted with DCM/Hexane (1:1)
Stock and Working Solutions are dissolved in 0.01M KH2PO4 buffer, pH 6.5

Is this change in RT significant? Is there any part in the sample preparation or Mobile Phase that can influence the change in RT?

I will appreciate any feedback!

Thank you.
Ghie Malig :wink:

Hello ghie

Good to see you are proceeding. Isn't the buffer molarity of your mobile phase very low 0.005 M or else 5 mM ...check this. I find it strange.

As stated by Srinivasan sir,

A variation in RT reflects the chromatographic parameters such as mobile phase composition (% B, pH ), temperature, flow rate, use of different column, solvents chemicals, different HPLC equipment (for ruggedness) affecting your separation. RT is a parameter to qualitative identification of analyte. Analyzing a single analyte may provide you wider range of varriation in RT but your method is less rugged and robust. A variation of +/- 0.1 min makes your method more robust and easily transferable to other labs (rugged) with reproducible results.

In case of impurity profile and stability studies a more strict control in variation of RT is needed.


Refer LCGC article over Retention time changes.

If you don't find this article let me know I will send you the link.

Good Luck!

Regards,

Amaryl.

Thanks Amaryl! The low molarity of the buffer solution in my mobile phase also caught my attention. :) I'll raise it up and see if there would be improvement with the shifting of RT. I really appreciate your feedback.

Best regards,
Ghie Malig :wink:
Hello ghie,

Good Morning! Here is your link over retention time changes.

http://www.lcgcmag.com/lcgc/data/articl ... rticle.pdf


Good Luck.

Regards,

Amaryl.

do you stir your mobile phase?
Once I had the problem, the mobil phase separated (invisible)
and so my retentiontimes shifted, run by run.
The good thing about a bowling future is that it starts with your next game, and comes only one game at a time

To Amaryl and to all those who replied to my query!

Thank you so much for answering all my queries. It will really be of help. Our next project will be beta blockers atenolol, propranolol and metoprolol. will also be developing method for diltiazem. All in plasma solutions. So I hope you will be able to help me again. I hope i'll also be of help to you in the future.Thank you!

Ghie Malig
7 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry