I agree with Don, the baseline drift is acceptable for the size of peaks that you have, provided that the integration puts the baseline in the right place as in Fig 4. Before you start swapping ferrules and trimming the column you need to check for leaks - if there is no leak there is no point swapping ferrules (do not fix what ain't broke rule), if there is a leak you must fix it before you trim the column. Check the septum also.

Peter
Peter Apps
I disconnected the column from the detector and inserted a no-hole ferrule. I ran my method and this is the chromatogram I got:


Image


Figure 6: Blank run-No injection-Nocolumn connected to detector. Temperature program: 170 deg C for 0 min, then 0.6 degC/min to 207.5 deg C for 0 min, then 5 deg C/min to 225 deg C for 0 min. Column: ZB wax 100% PEG.


So I guess it was the column and the detector does not need to be cleaned.
Thanks Peter and Don !!!
18 posts Page 2 of 2

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry