Hi HPLC experts,

Is there any relationship between fluorescence and UV detector for vitamin E determination in normal phase HPLC.
This problem arise because the fluorescence HPLC is currenctly using by another researcher (may take few weeks for her to complete) while UV HPLC is unused.So I planned to evaluate the efffect of few variables e.g. temp and time to the tocols in my sample using HPLC- UV/Vis. After that, I wish to use fluorescence HPLC for method optimization.
Can I follow this way??.
Will there be a difference in amount regardless the tocols concentration

PLease advise me.
Thank You
Our Standard method for determination of Tocopherols and Tocotrienols in oils uses UV detection, I have no idea about fluorescence.

However, if each method is calibrated correctly you should obtain similar results from the same sample from both assays .
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
2 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry