Clarification on OPGV

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
So I'm definitely a little confused about OPGV.

30-m column, 0.32 mm ID and 0.25 um film, carrier gas is helium

Various sources give different values for the OPGV of helium - some say 25 cm/s, some 30 cm/s and some 35 cm/s. According to the van Deemter plot, the optimal value would be of 27-28 cm/s for helium (although I think the OPGV is usually defined as being 1.5-2 times that value for ultimate compromise between resolution/speed)

In the Agilent manual, it is suggested to use a flow of 2.6 mL/min for a 0.32 mm column (this would lead to a head pressure of 11 psi and an average flow rate of 40 cm/s - NTP). In the same manual, they also suggest a pressure of 7 psi (which would lead to a column flow of 1.5 mL/min and an average flow rate of 25 cm/s!).

Now I'm aware that the OPGV value varies with temperature, along with several other factors/variables and I'm also aware that the relationship is not linear and that slight deviations from the OPGV should not affect the outcome of the chromatography by much). What I am looking for though is just a general good starting point to set in the default method I wish students to use to develop their own methods. What would be a good starting value for the average flow rate/column flow/head pressure...?


Thanks,

Roxanne.
Dear Roxanne,

I would like to start with some shocking news. OPGV (optimal practical gas velocity) of helium at 100 degree Celsius in 10-m long, 0.1-mm id (internal diameter) column is close to 90 cm/sec. For 1-m long column with the same id (these columns are popular in a certain class of real applications), OPGV is about 210 cm/sec.

I know that those are not the columns you are asking about. However, the examples provide a reality check for other widely known recommendations.

OPGV was proposed by Scott and Hazeldean in 1960 [1] as “average linear gas velocityâ€
Hi there,

Thank you so much for the reply. It does clarify things.

I also searched 6890 manual for the recommendation to use 7 psi pressure for your column. The only such recommendation that I found was for column conditioning. This is different from using a column for analyses of real samples. Please let me know if I miss other 7 psi entries.


It's in the 6850 Series Control Module User Information manual actually (p. 40), just below the recommended starting flows/pressures.

Again, thanks!

Roxanne.
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry