Anyone running EPA 548.1 for endothall?
Looking to bring this online at our lab instead of subbing it out. I was feeling encouraged until I noticed an AZ DHS info update claiming the extracts would quickly "ruin the column for other analyses".... yikes.
If you run this method - is this true?
The instrument currently runs low-level PAH in soil and groundwater, and we don't get enough endothall samples to justify having to constantly change the column.
I run EPA 548.1 and 525.2 on the same instrument. I do have to swap liners and trim a little column before switching back from 548 to 525 so I can pass the DDT/Endrin breakdown check. If you are only doing PAH and are not required to pass a breakdown check then it shouldn't be a problem.
Not being careful and leaving some of the acidic methanol in the extracts is what will ruin a column. If you have enough sensitivity on the instrument run them with a 10:1 split injection and it will keep the inlet much cleaner. Also using a guard column (1-2m) in front of the analytical column helps.
Thanks James - PAH is by 8270 so my only concern would be benzidine tailing. I use a drilled uniliner for that instrument, so maybe I could get away with a microinjection/short splitless hold.
Are you using an uncoated guard column?
Do you use BioRex cartridges for extraction? Phenomenex claims their Strata-x phase reduces ionic interference, but I'm always leery of promotional science.
We use the Restek cartridges
https://www.restek.com/catalog/view/1812/26063
They seem to work really well.
I normally use the Intermediate Polarity guard column from Restek, since I also inject ethyl acetate and it is a little too polar to give good peak shapes on the non-polar 5 phase column.
In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot] and 22 guests
