LC/MS instrument not present?
Posted: Fri Nov 05, 2004 11:14 am
We have been running some samples on a micromass triple quad mass spec (running masslynx 4.0).
The computer intermittantly loses contact with the instrument during a run.
The aquisition hangs and the software thinks it is still aquiring data. The LC side of the instrument carries on as normal.
The error seems to happen randomly, sometimes a run will be fine, other times it will "die" during an important sample.
Normally we would just ring waters (micromass), however it is only happening with this method.
The hard drive is not full.
We have tried pinging the MS after an error and the computer can't see it until we reboot
Is there anything else to check?
Specs:
The MS is attached to a waters 2695.
Flow rate is 1.3ml/min split 1:9 before entering the mass spec.
We are running 2 analogue channels (UV/Vis absorbance detector) and 2 MRMs. No scans.
Thanks
(confused) James
The computer intermittantly loses contact with the instrument during a run.
The aquisition hangs and the software thinks it is still aquiring data. The LC side of the instrument carries on as normal.
The error seems to happen randomly, sometimes a run will be fine, other times it will "die" during an important sample.
Normally we would just ring waters (micromass), however it is only happening with this method.
The hard drive is not full.
We have tried pinging the MS after an error and the computer can't see it until we reboot
Is there anything else to check?
Specs:
The MS is attached to a waters 2695.
Flow rate is 1.3ml/min split 1:9 before entering the mass spec.
We are running 2 analogue channels (UV/Vis absorbance detector) and 2 MRMs. No scans.
Thanks
(confused) James