EPA 515.4 surrogate issue

Posted: Fri Oct 18, 2019 1:10 pm
by Bigbear
I've been doing 515.4 for decades and once in a while a sample will fail the 30% acceptance range for 2,4-dcpaa. Spikes for these samples also fail the surrogate, but the 7 calibrated herbicides are all acceptable. The internal response for these samples are very close to the average of calibration. In these samples there is a very large peak about 4 minutes before the surrogate, but is not present in our method preserved blank water.
Anyone else see this? It seems as if 2,4-DCPAA does not track the other herbicides in the method.

Thanks

Re: EPA 515.4 surrogate issue

Posted: Thu May 06, 2021 5:45 pm
by chog
Not sure if you had any responses to this old one but we found the following issue with preservation in the method - more is not better. Other analytes were not affected.


Laboratory 515.4 Preservative Study
Table of % Recoveries of Surrogates and Spiked Analytes
Sample Type and mg of Na2SO3 Added
Austin tap DI Water
3 10 20 3 10 20
DCAA surrogate 96 51 33 101 55 31
2,4-DB 90 17 12 102 19 12
Chloramben 98 71 52 109 65 39