Understanding Thermo, understanding quadrupole

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
If you have a Thermo TSQ Quantum GC QQQ mass spectromter, some of their representitatives (some with Ph.D. education) should have told you when you input the mass value, you must be accurate to the second decimal (eg. m/z 118.88, not 119, nor 118.9), and use a mass scan window of 0.01 amu or less, or you lose both sensitivity and stability. This claim is, without doubt, beyond the resolution power of a classic quadrupole.

However, one person in my lab had some tests, and did find signal increase. Is it just a random coincidence, or faith works? but, it also means you may have to update your mass everytime you run the method.

Any opinion on this?

My lab got a TSQ Quantum Discovery. The engineer only told me to set the mass ion to one decimal place. Scan window is set down to 0.01 Da for better specificity, if necessary. Nothing to deal with sensitivity.
2 posts Page 1 of 1

Who is online

In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry