extraction of docetaxel from lactated ringers for LC/MS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
4 posts
Page 1 of 1
Hi guys I am new to LC/MS and heard this is the place to get answers. I am trying to detect docetaxel/paclitaxel in a sample of lactated ringers. The lactated ringer's is a solution of electrolytes (NaCl, KCl, Ca, lactate). I have been getting ion suppression of the docetaxel peak. I am trying to use solid phase extraction but I cannot seem to get the ion suppression/ electrolytes to go away. Has anybody had experience with lactated ringers or a better desalting extraction?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
On a reversed-phase column or even reversed-phase SPE, the ions should separate readily from the analytes. What are your chromatography conditions?
-
- Posts: 2
- Joined: Wed Feb 18, 2009 5:48 pm
Chromatography and detection:
-waters 2695 hplc
-micromass zq2000 single quad MS using positive ESI; 3.5kV cap voltage; 25V cone; 120C source temp; 450 desolvation temp
-quantifying the sodium adduct since it was the biggest peak during tuning (830.25 m/z)
-waters xterra c18 3.5um 2.1x100mm colum/ guard column 2.1x20mm 3.5um c18
-mobile phase 50% 40uM sodium acetate with 3.4 mM acetic acid/ 50% ACN; added sodium to force only 1 adduct formation rather than K or NH
-isocratic flow rate 0.2 ml/min
Extraction:
-discovery 96 well SPE plate with 50 mg C18
-method is still being changed using different wash solvents/volumes
-???
I think the problem is that the electrolyte concentrations are very high in lactated ringers compared to my drug concentrations that I spike. mg/ml vs. ng/ml
-waters 2695 hplc
-micromass zq2000 single quad MS using positive ESI; 3.5kV cap voltage; 25V cone; 120C source temp; 450 desolvation temp
-quantifying the sodium adduct since it was the biggest peak during tuning (830.25 m/z)
-waters xterra c18 3.5um 2.1x100mm colum/ guard column 2.1x20mm 3.5um c18
-mobile phase 50% 40uM sodium acetate with 3.4 mM acetic acid/ 50% ACN; added sodium to force only 1 adduct formation rather than K or NH
-isocratic flow rate 0.2 ml/min
Extraction:
-discovery 96 well SPE plate with 50 mg C18
-method is still being changed using different wash solvents/volumes
-???
I think the problem is that the electrolyte concentrations are very high in lactated ringers compared to my drug concentrations that I spike. mg/ml vs. ng/ml
-
- Posts: 27
- Joined: Tue Aug 15, 2006 11:11 pm
Isocratic conditions is likely to be your problem.
Switch to gradient conditions and allow for a brief period of high aqueous at the beginning of your run to wash out all of the salts with a quick ramp to high organic (methanol).
Switch to gradient conditions and allow for a brief period of high aqueous at the beginning of your run to wash out all of the salts with a quick ramp to high organic (methanol).
4 posts
Page 1 of 1
Who is online
In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 21 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 21 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science