It has been suggested in previous posts about similar problems, that this might be due to a blocked split line. How can I actually clean the split line?
Column flow is 2.5 mL/min (constant flow mode), pressure 23 psi and the split is 40, but the problem persists with higher (5.0 mL/min) or lower (1.0 mL/min) flow rates. When I turn the split off, the pressure stabilizes within 10-15 sec and the GC switches into ready mode. If I then turn up the split (changing and reactivating the method) the pressure remains constant up to a split of 30. When I then increase the split to 40 the pressure becomes unstable (fluctuating between 20 and 25 psi).
I measured the flow (rotameter) at the split vent and the flow is lower than I would expect (?). E.g. at split 30 the reading at the instrument sofware is 82 mL/min total flow, the actual flow at the spilt vent is approx 40 mL/min. At split 40 the pressure starts oszillating and the flow at the split valve starts to fluctuate between nil and burst of 100 mL/min.
The GC has been running with this method before and the problem seems to have started after the injector was cleaned. This leads me to believe that there might be are particle in the split line. How can I check that and how can I clean the split line?
Any help will be highly appreciated, as I am getting kind of desperate...