Page 1 of 1

waters 991 PDA

Posted: Wed Jan 28, 2009 4:30 pm
by jongf
I am looking for software for the 991 pda.
Also data cable and manual would be helpful.
Thanks

Posted: Wed Jan 28, 2009 4:56 pm
by AA
The 991 has not been supported for a number of years (like 10) so users guides could be hard to come by. If I remember correctly it uses a IEEE cable to control. As for software, you are looking at finding an old version of Millenium. You may be able to get a signal out of the analog output but I think if you want it to work as a true PDA with control you may be out of luck.

Posted: Wed Jan 28, 2009 8:22 pm
by juddc
I might have a manual for that...and it should run on an old Millennium system. The cable is an IEEE-488, I think. If I recall correctly, it's not the most sensitive detector - the 996 is about an order of magnitude quieter and I'd recommend one of those highly. With that said, the 991M was a reliable machine.