Hello,

I'm using a Shimadzu 2010 GC-System with an OCI wich runs with glas-liner in Split-Mode at a ZB5 Column. The Glas-Liner are filled with glas-wool. The problem is that the wool sliped down the liner and blocks the flow onto the column and the result is an unuable system. The Shimadzu-Service can't give me an answer why the wool slip down. Does anybody else have that problem with this liner? The Shimadzu Part.Nr. is P/N961-01481-02
Did this just start happening, or is this the first time you've tried? Are you inserting it with the correct orientation? Wool should be more towards the bottom; needle shouldn't touch it like with focus/precision liners.

Are there large pressure/flow changes in the inlet in the method? We had that happening once with our Thermo PTV in split mode (same basic idea design as the Shimadzu OCI/PTV) and remedied it by changing inlet parameters.

Can you try a liner from a different vendor? Restek sells ones for the 2010 OCI/PTV
2 posts Page 1 of 1

Who is online

In total there are 38 users online :: 1 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: Google [Bot] and 37 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry