Chemstation sequence saving PDF

Posted: Wed Oct 09, 2019 4:56 pm
by Chemiger
Hi Everyone,
I am using the set up listed below, I am having an issue that every time I am running a sequence it runs a single sample then needs me to name the pdf copy of the sample file then save it before it moves on to the next sample in the sequence, does anyone know of a way to bypass this feature ?

GC Agilent 6890N (G1530N)
MS Agilient 5973 network Mass detector G27579A
Autosampler 7683
Software
Agilient MSD Productivity Chemstation for GC and GC/MS systems. Data analysis application
(MSD Chemstation G1701EA E.02.02.1431)
Thanks,
Chemiger

Re: Chemstation sequence saving PDF

Posted: Wed Oct 09, 2019 7:15 pm
by James_Ball
It is because you have your printer listed as printing to PDF and it requires a unique name for each time it prints. It also means you have it automatically running Data Analysis at the end of each analysis run. You can either send it to a regular printer or turn off the Data Analysis with each run by unchecking the Data Analysis check box in the Run Method screen that pops up before you begin the sequence.