Advertisement

The intensity test on UV detector (HPLC 1100 agilent) felt

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Any body knows why the intensity test on UV detector felt. I have changed the cell and the lampe ( the new one).

thanks

Go to Diagnostics on the software, there is a self-test.

Also, check the basics.
Perform the cell test ( which will tell you if it's the cell ) - ensure you have standard cells, the low volume cells will fail, as may an empty cell - cell needs to be filled with water. If the failure is at low wavelength - 190, could be lamp. Newone may not have warmed up fully - give then at least 30 minutes

Bruce Hamilton.

.
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry