vliu » Thu Jan 08, 2009 10:21 pm
Danko:
The high result refers to both height and area.
Just to give you a clearer idea, below is the before and after result
result before cleaning:
(area; k prime; height; symmetry, plate count USP)
8180353; 4.54; 329829; 1.75; 8357
result after cleaning:
9780618;4.38;433186;1.66;10127
Uwe Neue:
Quantitate by peak area.
My results will be compare to another lab, and their results are comparable to my old result (their column is newer than ours). I am getting so confused. I have been cleaning the column (without connecting to the detector) overnight with 50% acetonitrile. My plan for today is to give the system inclu the PDA detector a good clean with;
1. warm water, both lines
2. 6M HNO3
3. warm water to rinse out all HNO3
4. reintroduce my mobile phase and do a trial injection
if still doesn't work then......I don't know what else I can do......