Interesting question, because I stopped my previous response in talking about since it was a not quite what you were asking about. In fact though, one of the problems with how the inlet is configured is it impacts using a purge & trap on the instrument and probably also a headspace unit (I've not set one up, but I would suspect that they're interfaced to the system the same way as a purge & trap).
If you think about how the inlet works, at all times, we have (using my previous example) 104 mLs/min going into the inlet. We switch from split to splitless by switching the flow from going down through the inlet (split mode) to across the top of the inlet (splitless mode). Now, with a syringe injection, the sample does not "see" the flow across the top so we get most of the injection deposited into our column. The problem arises when we are introducing the sample as part of the flow into the inlet. When we hook up a purge and trap to the GC, we typically take the flow from the GC to the back of the purge and trap, and then hook the purge and trap up to the inlet where the carrier gas normally enters the inlet. So, with how our inlet works, we're effectively always in the split mode regardless of what the valve is doing. The analytes that are coming in are simply going to follow the flow, even when we're in "splitless" they'll simply flow out the septum purge line along with everything else.
So, what to do?
1) You can hook the transfer line up directly to the column (there are panels on the top of the 5890 GC you can unscrew that allows you to pass the line through). You can then use the mass-flow controller to maintain the flow. The problem that is common with this configuration is the peak shapes of the early volatiles, particularly the gases.
2) You use Greg's trick. The predecessor of the 5890 was a GC called a 5880 (I'm dating myself here, though I also used the 5880's predecessors), which had a very similar inlet, but when you were in the splitless mode, the flow to the inlet was only what was going into the column (and also out the septum purge). The problem HP (now Agilent) had was flashback. The liners they used with the instrument were very narrow, and with the very low flow, you had a lot of material flashing up into the top of the inlet and actually the carrier gas tubing. Now, for volatiles we're not making a liquid injection so that's not a problem. the only significant difference between the inlet on the 5880 and the 5890 was how the inlet was plumbed.
This diagram is how your 5890 is most likely plumbed:
You need to convert it to:
(I was in the midst of trying to use words to explain this and came to the conclusion that it was going to be much easier to draw a diagram.)
What this configuration does, is that when the instrument is in splitless, the only flow entering the inlet is going on column or out the septum purge line. The flow that ends up going out the split vent is diverted around the inlet.
Assuming you're using a megabore type column, you should have sufficient flow for desorbing the trap. Typically I would like at least 10mLs/min to desorb the trap.
The benefits, are that you should get much better peak shape for the gases, and when the inlet switches to split, the increased flow will work to sweep moisture and heavier organics out your split vent.
In doing this, if you're unfamiliar with how the lines are hooked up in a 5890, Agilent (or HP) uses connectors that are plastic (usually white) nuts that need to be only finger-tightened. However, they use a small rubber o-ring which has the annoying habit of being stuck in the hole when you unscrew the line, or rolling off to hide place under the instrument so that when you put it back together you have a massive leak. My point is simply, double-check that you have only one o-ring (that one isn't stuck in the hole) and or if you think it's stayed in the hole, it didn't just roll away and hide under your instrument.
Take a look at the diagrams and feel free to email me back to better understand what I'd trying to explain.