High split ratio alternative

Posted: Fri Sep 20, 2019 4:51 pm
by Phosgina
Task:
Switch to N2 carrier gas

Current practices:
He carrier gas, backpressure-regulated GC's (Agilent). A decent amount of work is injecting neat solvents for purity, use 400:1 split ratio to prevent overload. He is limited by EPC to a total flow of 1250mL/min.

Challenge:
N2 is limited by EPC to a total flow of 200mL/min. The 400:1 split ratio isn't possible.

Any suggestions?

Re: High split ratio alternative

Posted: Fri Sep 20, 2019 4:56 pm
by Consumer Products Guy
Dilute the sample.

Re: High split ratio alternative

Posted: Mon Sep 23, 2019 12:03 pm
by GasMan
Are your impurities overloading the column? If not, you could try a splitless injection to get rid of the solvent tail.

Gasman

Re: High split ratio alternative

Posted: Mon Sep 23, 2019 2:55 pm
by James_Ball
What detector are you using?

What are the dimensions of the column you are using?

Different options can be found depending on more information.

Re: High split ratio alternative

Posted: Tue Sep 24, 2019 11:22 pm
by AICMM
Phosgina,

what is keeping you from using H2 instead of N2?

Best regards,

AICMM